    TESTING OF BONES

    On 28th August 1992, I offered Moshe some advice regarding his
    proposed testing of rat bones and I note that he has decided to use
    a bending test with the PMMA potting.

    I guess it would have been helpful to provide my comments to the
    rest of the network; however, I thought that most of these items
    were well understood in the biomechanics community - or, perhaps,
    that's why Moshe obtained only a few replies. Consequently, I
    include my original comments below and provide some additional
    comments.

    ORIGINAL COMMENTS

    As with most bones, you will find that the mechanical properties
    are dependent upon:

    + Location of the sample.
    + Species.
    + Age.
    + Exercise.
    + Individual tested.
    + Other possible factors.

    Destructive tests (comparing a treatment with some series of
    "controls"), therefore, are fraught with the statistical challenges
    of these biological variations. Any studies that cast aside
    concern for these variables are likely to produce such large
    standard deviations that no meaningful differences can be
    discerned.

    Ultrasound (as a non-destructive tool) offers great advantages for
    such monitoring; however, an "error analysis" should be conducted
    (using the known dimensions etc and their associated measurement
    errors), to ensure that the measurement tolerances will not
    overwhelm the anticipated differences that are to be measured. The
    size of the rat bone may well be a dominant factor in whether you
    can expect useful data and the Error Analysis will advise you of
    this problem BEFORE you start to sacrifice animals and time.

    The non-uniform shape and anisotropic properties of long-bones tend
    to frustrate the repeatability of bending tests; consequently, many
    researchers have resorted to torsional tests. You may still wish
    to use a guide-wire (surgical k-wire) to establish a meaningful (?)
    axis for the bone.

    While some researchers use low-temperature bismuth alloys for
    potting their samples, these materials can be quite expensive.
    Medical researchers often use PMMA due to its availability in the
    hospital - material is often available free in the hospital, after
    expiry of a "shelf life" or the opening of a larger pack that has
    not been completely used. Bulk purchases of PMMA from Dental
    Suppliers may be cheaper than the material used for total joint
    replacements. Due to the large amount of heat given off by curing
    (polymerizing) PMMA, the volume of PMMA should be kept to a
    minimum. 5-minute epoxy is often readily available to non-medical
    workers and is relatively cheap. Any of these techniques are
    suitable for potting, so long as attention is paid to minimizing
    potential thermal damage to the specimen.

    ADDITIONAL COMMENTS

    I note that Moshe has decided to use a Bending Test with an Instron
    machine and has encountered loosening problems with the samples
    potted in PMMA.

    INSTRON: It is unclear whether the chosen machine is screw-driven
    or electro-hydraulically controlled. Due to the smallness of the
    samples (rat bone), a screw-driven machine could present additional
    problems in obtaining smooth curves. No matter whether the machine
    is screw-driven or electro-hydraulically controlled, the resolution
    of the position-measurement and force-measurement transducers will
    be of particular concern.

    BENDING TEST: It is still unclear whether the proposed tests are
    destructive (with all of the inherent problems listed above) or
    whether the same sample is to retested within its "elastic" range
    after a variety of treatments.

    A bending test will require some mechanism for controlling the
    axis-of-bending in relation to the principal axes (i.e.
    geometrically-defined - second moment of area) of each bone.
    Attempts to force bending to occur on any other axis than the
    "minimum" principal axis may well result in the bone trying, during
    loading, to rotate on the loading and support noses.

    ASTM provides guidelines (D790M-86 standard on plastics) for span-
    to-depth ratios for bending tests on machined samples of materials.
    For anisotropic materials, span-to-depth ratios of 60:1 are
    recommended! Where samples may be of various dimensions, the
    failure to control "shear effects" (by adhering to an appropriate
    span-to-depth ratio) may create additional problems and increase
    the variance in the data.

    ASTM-D790M-86 also provides guidelines for the radii of the loading
    and support "noses", "... to avoid excessive indentation, or
    failure due to stress concentration directly under the loading nose
    or noses..."

    ERROR ANALYSES: An Error Analysis should be conducted BEFORE any
    type of testing and is not just a requirement for the Ultrasound
    Testing mentioned above.

    DEHYDRATION: Moshe does not mention any precautions that he is
    taking to control the extent of dehydration of the bone.

    The curing PMMA alone will generate heat that can dehydrate the
    bone. The significance of such dehydration is not widely
    published. In a recent paper:

    Finlay et al, 1992: "Embalming effects upon the mechanical
    properties of bone: Preliminary experiments" In "Experimental
    Mechanics: Technology transfer between High Tech engineering and
    biomechanics", Little, E.G. editor, Elsevier Science Publishers
    B.V., Amsterdam, 185-200,

    (Details of the proceedings of this conference were provided on
    BIOMCH-L on 19th September 1992)

    we reported briefly upon the dramatic anisotropic contractions that
    can be seen when the cortical bone of a bovine femur is dehydrated.
    The contractions were 0.6% axially, 2% circumferentially, and 3%
    radially and were statistically different from one another. These
    values translate to strains of 6,000, 20,000 and 30,000 microstrain
    respectively - i.e. considerably above the 3,000 microstrain peaks
    that are reported in the literature for normal activities of daily
    living! While embalming reduced these strains, they still
    remained at levels of 0.5%, 1.2% and 2.1% respectively (these are
    all approximate figures without quotations of the standard
    deviations).

    Rehydration of the bovine samples in our tests brought the axial
    dimensions back close to their original values; however, the
    circumferential and radial dimensions remained notably different
    from their original dimensions.

    These findings for the "contractions due to dehydration" validate
    observations quoted to me by Dr Sidney Lees of the Forsyth Dental
    Center in Boston. (Sidney is not on e-mail). Clearly this
    information on "dehydration/ rehydration" strains needs to be
    disseminated more widely - any editors out there who would like to
    review such a manuscript for their journal?

    Obviously, any contraction effects due to dehydration of the rat
    bone will add to the contractions encountered during polymerization
    of the PMMA and will produced a loosely-potted sample.

    Careful choice of a bismuth-alloy will provide an alloy that
    maintains its dimensions during cooling; however, it will be more
    expensive than PMMA. The volume required for testing rat bones
    should, however, be quite small.

    To aid in the alignment of long-bones during potting, we regularly
    use screws (3 or 4) in each pot. These screws and each pot can be
    coated with Vaseline (or some other "thin" grease), to permit easy
    removal and reassembly of each bone in a single set of pots. In
    this way multiple sets of pots are not required. The screws may be
    utilized to "take up" any "small" amounts of "play" between the
    bone and its pot; however, we have not encountered this "looseness"
    problem with bones from humans, cows or sheep.

    I hope these additional comments are of value and I would encourage
    other BIOMCH-L members to post their comments for general viewing -
    obviously, I was in error by not displaying my original comments to
    Moshe.