    ORIGINAL QUESTION:
    > I am using a portable EMG system and utilizing 6 of 8
    > available channels for data collection. Usually only 60hz is emitted by
    > electrical devices, but I am finding out that computer monitors are
    causing
    > more than just 60Hz interference--they are causing the EMG system to pick

    > up interference at approximately 60-62Hz, 120-123Hz and 180-184Hz during
    a
    > low exertion VDT task. I have tried various computer monitors
    > and they all seem to cause this "excessive" interference. When I turn
    the
    > monitor off, only 60Hz remains (from other equipment). Has anyone run
    into
    > a problem like this? If so, what has been determined as the cause and
    what
    > could be done to remedy this situation?

    My thanks to those who took the time to provide very helpful suggestions.
    An LCD Display took care of my noise problem.

    Mike Monroe




    RESPONSES:

    From:
    Hartmut Witte
    In a comparable situation we found that our force measuring chain (special
    made
    Kistler plates - amplifiers - ADC - Computer) owned a single peak in the
    basic
    noise level at 78 Hz (cf. attached file noise.doc). This turned out to be
    the
    frequency of the graphics card and the monitor. We minimized the
    interferences
    by testing several combinations of graphic interfaces, their position on
    the
    main board and monitors, but could not eliminate it, but reduce it to about
    the
    tenfold amplitude of the base noise level.
    We remain as long as possible in the frequency domain, apply NO filtering
    and
    check the "last" data in the freq. domain for resonance effects. Up to now
    (knock on wood!), there was no need for us to filter or smooth before
    changing
    into the time domain. In EMG-data, this for sure will be different. Perhaps
    you
    may use your two unused channels to control the noise level (freq.
    dependant)?
    If you find any graphic interface which does not use the main board as an
    antenna, please tell me!



    From:
    Dave Grimshire

    As you know computer monitors radiate electro-magnetic energy. The human
    body is a very good antenna and collects the energy as you've seen on your
    EMG. There are low radiation monitors made that should reduce the amount
    of energy a monitor will direct to the front of the screen. Also small
    monochrome monitors seem to have less radiation.

    Dave Grimshire

    From:
    Jan Cabri jcabri@zatopek.fmh.utl.pt

    Why don't you try a portable computer (laptop). If you do not
    connect the laptop to the power line, I think it might resolve your
    problem (I guess your electrodes are not shielded enough to prevent such
    an interference). However, you will have to buy a pcmcia-a/d convertor
    for this, which don't go cheap.
    Best regards,

    Jan Cabri, Ph.D.
    Technical University Lisbon

    From:
    Gonzalez
    It's normal to find components at integer multiples of the frequency
    (60Hz) of the periodic signal. They are called harmonics and they will be
    present as long as the periodic signal is not a perfect sinusoid.
    What can be done? To turn off any equipment not involved in the
    measurement, including lights. Also, apply some type of filtering after
    acquisition of the signals, either analogue or digital filtering. Filters
    used for supressing this type of interference are known as notch filters.
    Tradeoff: The part of the signal (EMG) near 60Hz, 120Hz, 180Hz, etc will
    be affected as well.

    Regards,

    Jose A.


    From:
    "Murray Maitland"

    You might wish to look at your pre-amplifiers. Generally the
    closer the preamps are to the signal source the better in terms
    of increasing the signal to noise ratio.

    You might look at using a notch filter to remove the 60 Hz
    and harmonics. See The Scientist and Engineer's
    Guide to Digital Signal Processing
    by Steven W. Smith
    California Technical Publishing
    ISBN 0-9660176-3-3 (1997)
    http://www.dspguide.com/pdfbook.htm

    You might also look at your grounding since this can make noisy
    loops. Although for surface EMG, it isn't usually a big problem.


    From:
    Gordon Chalmers
    Here is a suggestion of various things to try. Have you tried a grounded
    shield around the top, sides and back of the monitor? Try a tinfoil sheet
    as a shield with aligator clip attached and run to ground contact on a
    piece of equipment (and allow air space for ventilation). Also, can the
    subject and subject keyboard equipment be physically and electrically
    isolated from the data collection equipment? (e.g., not on same table top,
    and/or large physical distance such as opposite ends of the room, perhaps
    even try different electrical circuits in the room for research monitor &
    equipment and the subject keyboard equipment, and the research subject and
    his/her equipment could be placed on rubber mats to isolate).

    Finally, since you know it is the monitor, and if nothing else works, you
    could try a flat screen monitor (or perhaps this is one of the other
    monitors you have tried), my guess is they would not emit strong electrical

    fields as CRTs do.

    From:
    "Arnel Aguinaldo"
    You are probably picking up the refresh rates of the monitors and/or the
    processing cards of the PCs, all of which have the capacity to run
    processes
    of up to 180Hz, depending on the system. My experience has shown that
    increasing sampling frequency and adding bandpass filters usually would do
    the trick. Otherwise, I'd try other signal conditioning methods.



    From:
    "Peter Chupity"
    Have you considered using an LCD display?


    From:
    Jason Harrison

    You're measuring harmonics. The refresh frequency of the monitor is
    probably 60Hz, then you have the horizontal frequency (much higher)
    and the vertical sweep frequencies.

    If you've already sheilded all your cables then you can try moving the
    monitor or using a projection monitor to move the monitor as far away
    as possible. Or use a notch filter to remove the frequencies from
    your measurements.


    From:
    Matthew Pepper

    Here are some ideas.

    You are picking up harmonics of 60Hz. Probably generated by the power
    supply.
    This is likely if the supply uses rectification to transform the AC wave
    form
    to DC. The rectification process produces the harmonics. The transformer
    will
    also produce harmonics as it is not a linear system.

    It may be that a switch mode power supply will help if the line
    interference is
    generated by your system. However if the monitor is the source its' power
    supply will already be switch mode.

    Otherwise the source can be other line powered equipment in the same lab as

    well as any fluorescent lighting.

    If the computer monitor is the problem you could try replacing it with a
    LCD
    display. If your system is flexible enough you could use a PCMCIA data
    acquisition card [$400] and run the experiment on a laptop.

    Regards

    Matthew Pepper

    From:
    Einar Skavland Idsø
    The signals at 60-120 Hz are the vertical screen updates (refresh rates).
    On
    current monitors a ray of electrons is fired at the monitor 60-120 times
    per
    second. Older monitors (and video cards) can only use the lower frequencies

    (