    I think that Scott's message highlights the biggest problem faced with all marker sets: skin movement artifact.

    When we compared bone-pin mounted marker clusters with skin mounted marker clusters, even when using the same anatomical coordinate system, skin mounted markers not only predicted larger motions, they often predicted motions that were opposite to what the underlying bones of the tibio-femoral joint during walking (1) and cutting motions (unpublished results).

    When we further evaluated the skin marker data and applied various "solidification" and optimisation methods to the marker clusters we had no improvement in terms of reducing the error between skin and pin mounted knee joint kinematics (2).

    In terms of a bias imposed by the skin movement artifact, although we attempted to find correlations across subject, the error was subject specific (unpublished results) and we have, as of yet, not found a reliable way to reduce this subject-specific bias.

    As such, we have attempted to put forward a basic guideline for reporting kinematics along the lines of what Scott was suggesting (3) by describing the standard error associated with predicting tibio-femoral motion from a four marker cluster of skin markers similar to that described by Manal et al (4). Although this was a conservative way of predicting the error, in terms of translations and abd-add and internal-external rotations the 'window' of error may exceed the actual measurement. This is particularly significant since skin markers may predict a tibio-femoral motion that is not only different in magnitude but also in direction to the in vivo motion of the underlying bones.

    Ironically, based on some of our other data I believe it may be easier to 'subtract' the error during more vigorous activities (cutting and hopping) due to the increased level of muscle activation which thus reduces the 'wobbling mass' effect of the soft tissue of the thigh (we are still looking at this however).

    In summary, I concur with Scott that we must consider the limitations of our technology, or perhaps even more so of our methodologies. Although the technology is there and it is so tempting to look at those graphs displaying the secondary motions of the joints (in particular the knee joint), we have a responsibility to ensure that the information we deliver respects those limitations or, at the very least, communicates those limitations within the paper. This is particularly relevant when publishing in journals that have a more clinical slant as the limitations may not be as apparent to this audience.

    Respectfully,

    Daniel Benoit

    References:
    (1) Benoit DL, Ramsey DK, Lamontagne M, Xu L, Wretenberg P, Renstrom P.
    Effect of skin movement artifact on knee kinematics during gait and cutting
    motions measured in vivo. Gait Posture. 2006 Oct;24(2):152-64.

    (2) Benoit DL. Motion analysis of the Knee: Kinematic Artifacts...PhD Thesis, Karolinska Institututet, Stockholm, Sweden. 2005 ISBN 91-7140-357-4.

    (3) Benoit DL, Ramsey DK, Lamontagne M, Xu L, Wretenberg P, Renstrom P.
    In vivo knee kinematics during gait reveals new rotation profiles and smaller
    translations. Clin Orthop Relat Res. 2007 Jan;454:81-8.

    (4) Manal K, McClay I, Stanhope S, Richards J, Galinat B.
    Comparison of surface mounted markers and attachment methods in estimating tibial
    rotations during walking: an in vivo study. Gait Posture. 2000 Feb;11(1):38-45.

    Daniel L Benoit, Ph.D.
    dbenoit@uottawa.ca
    Tel / Tél: (613) 562-5800, ext. 8547
    Fax: (613) 562-5428

    School of Rehabilitation Sciences / École des sciences de la réadaptation
    University of Ottawa / Université d'Ottawa
    451 Smyth Rd.
    Ottawa ON
    K1H 8M5 Canada-----Original Message-----
    From: * Biomechanics and Movement Science listserver [mailto:BIOMCH-L@NIC.SURFNET.NL] On Behalf Of Scott Tashman
    Sent: July 6, 2007 17:13
    To: BIOMCH-L@NIC.SURFNET.NL
    Subject: Re: [BIOMCH-L] Marker-set independent gait analysis

    To all who have contributed so far - I believe that this is a
    valuable discussion of an area too often neglected in motion
    analysis, and the quality of the responses has been excellent.

    I am not going to add to the discussion about the relative merits of
    different marker systems, but I would like to comment on how these
    systems are being evaluated/compared. In particular, I would like to
    emphasize one point (previously made by David Smith): Repeatability
    is necessary but not sufficient to establish accuracy. It implies
    good precision, but cannot address the issue of measurement bias.
    If one considers the nature of errors introduced by skin markers, the
    potential for biased measurements is significant. Displacements of
    surface markers due to muscle contraction, skin motion due to change
    of joint angle or tissue "bounce" due to impact forces (e.g.
    heelstrike) are not random - they are highly correlated with movement
    patterns. The errors introduced by these displacements can be quite
    repeatable within each subject, but can introduce significant bias
    that varies from subject to subject (depending on body style, injury/
    disease state, etc). We can generally agree that reduced within-
    subject variability (across trials/test sessions/investigators) is a
    "good thing". However, in the absence of a gold standard for
    comparison, we simply do not know how accurate our measurements
    really are. This has been an achilles heel of 3D motion analysis
    throughout its history. I remember working with Jim Gage at the
    Newington Children's Hospital gait lab back in the early-mid 1980's
    (using a marker system that might be considered a predecessor to the
    Helen Hayes system). We could generate repeatable knee internal/
    external and ab/adduction plots for our patients (mostly children
    with cerebral palsy), but ended up removing them from our clinical
    printouts because we did not feel confident about their accuracy. I
    routinely see these plots now; clearly the motion analysis technology
    has improved, but how much more do we know about their accuracy than
    we did 20+ years ago?

    So, I think the focus of the relative repeatability of different
    marker systems misses the point to some extent. It would be far more
    valuable to understand the effect of marker sets on the "confidence
    interval" of the measurements we use for research and clinical
    decision-making, particularly as they apply to the specific motions
    and subject populations each of us work with. It is simplistic to
    assume that there is one marker set that is the "best", even if we
    consider only clinical gait analysis. For example, a functional
    calibration approach (such as that described by Richard Baker) could
    be the best solution for a relatively healthy population, but might
    be poorly suited for subjects with limited range of motion in one or
    more joints (as suggested by Ton van den Bogert ,e.g. patients with
    CP, stroke, arthritis, etc). Landmark-based joint center location
    may work better for these populations, but probably has larger errors
    with obesity or skeletal deformity. The type of movement task also
    changes the nature of errors (consider skin displacement during
    running vs. slow walking), and might affect the relative performance
    of different systems.

    Trying to summarize where I am going with this (I almost forgot by
    now), I think the search for the "best" marker set is a bit of a
    quest for the holy grail. The obvious weaknesses of skin-based
    motion measurement for assessing joint kinematics have been
    demonstrated in multiple studies, and no marker placement or analysis
    method can completely overcome them. Perhaps the best we could hope
    for would be to agree that there are multiple approaches, each with
    their own merits. They should be selected based on consideration of
    the specific application, rather than strictly by convenience ("came
    with the system" or "that is the one I know"). Equally important, we
    should consider the inherent limitations of this technology, to avoid
    interpreting artifact as science (even if it is repeatable). Many
    types of studies may be relatively insensitive to marker set
    selection. For others, there may be specific methods that are
    clearly advantageous. However, for studies of certain disorders and
    subject populations, there simply may be no acceptable surface marker
    solution.

    Then, how do we answer the original question about marker set
    selection? I do not think we have the data to make these decisions.
    But, I believe that one key step is incorporating true studies of
    accuracy (instead of just repeatability) into the decision-making
    process. This is an area where emerging technologies (e.g. dynamic
    MRI or biplane radiography) have the potential to make significant
    contributions to our understanding of the estimation of skeletal
    motion from skin markers.

    Thanks to Nancy Denniston for initiating this topic, and I look
    forward to continued discussion/debate.

    Scott Tashman