  • Real-Time Latency of VICON

    Dear All Biomech-L subscribers,

    The issue of real time latency from Vicon in particular, and motion
    capture systems in general has been raised on this list recently.

    To (hopefully) clarify the situation, I would like to explain what
    contributes to latency, and how the data flows through the Vicon system.
    I assume other passive marker-based optical motion capture systems are
    similar, however the below analysis applies to Vicon MX systems only.

    This is the chain of events:

    1) A number of markers are moving around in the calibrated Vicon space.
    The strobe connected to the MX camera will emit light, which will
    reflect off the markers and be projected back onto the digital sensor
    inside the camera. The freeze-frame shutter used by the Vicon cameras
    will open to co-incide with the strobe light. For low frame rates (up to
    approx. 120Hz) the shutter is open for 1ms, but as the frame rate
    increases the shutter time (and strobe time) drops according to the
    formula t (in ms) = 120 Hz / frame rate (in Hz).

    2) As soon as the shutter closes, the camera will scan the sensor. A
    digital sensor is scanned line by line, and this takes some time. The
    Vicon Vegas 4 mega-pixel sensor used in MX-F40s is able to scan all 4
    mega-pixels in 2.4ms, whereas the sensor used by the previous generation
    MX40 takes 5ms to scan the sensor. At 1000Hz, the sensor is only
    partially scanned, and the scanning takes 1ms minus the time the shutter
    is open at that speed, i.e. 1ms - 0.12ms = 0.88ms (obviously it can't
    take longer).

    3) The data is then circle fitted and packaged up for transmission to
    the PC via ethernet. The time this takes depends on the amount of data
    to process, but for low marker counts (say 10-40) it is less than 0.5ms.

    4) The PC then has to gather all ethernet packages from all cameras. If
    the amount of data is limited, say 10-40 markers, this takes
    microseconds since the cameras send centroid (x,y) data only.

    5) The PC then reconstructs and labels all the camera data. Clearly, the
    time this takes is heavily dependent on the number of cameras, the
    number of markers, the power of the PC's CPU(s) and the complexity of
    the labelling problem. However, again working with typical marker sets
    of 10-40 markers, experiments have indicated that this takes on average
    1ms to 5ms.

    6) Finally, the PC does something with the data. For example, it can
    render to the screen, generate a sound or control a device. Rendering to
    screen or projector introduces significant latency, depending on the
    update frequency of the screen and how the data is rendered, but this
    could well add 10-30ms to the latency, and is largely dependent on other
    factors than those that can be controlled by the manufacturer of the
    motion capture system.

    It is worth noting that performance will depend on the number of other
    tasks the Windows operating system performs. It is therefore important
    to avoid as much as possible running other applications and services as
    these may cause increased average latency.

    The above figures should give the correct ball park: at 1000Hz the
    latency on the camera + transmission to the PC should be in the region
    of 1.5ms, at 100Hz in the region of 3ms. In addition there is the
    processing that takes place on the PC, which is dependent on the PC
    itself, the number of markers/cameras, the complexity of the labelling
    problem and other processes/services that compete for the CPU time. This
    has experimentally been shown to be between 1 and 5 ms for commonly used
    marker sets between 10 and 40 markers.

    Comparing passive optical systems to active marker systems is also
    interesting. Active marker systems have one major advantage and one
    major disadvantage when it comes to latency. The advantage is that
    active systems do not have to do real time labelling of the
    reconstructed 3D points since only one marker is visible at the time.
    The disadvantage is that because only one marker can be visible at the
    time, the latency will scale linearly with the marker count. For
    example, if the active system tracks at 5000Hz, each marker will add
    0.2ms to the inherent measurement latency, so 10 markers will take 2ms,
    20 markers 4ms, etc.

    Best regards,

    Lasse Roren
    Product Manager Life Sciences
    Vicon