  • Summary: 3DGA ankle marker placement for prosthetic feet

    Dear all,

    Please find the below summary of responses from my questions on marker
    placement of prosthetic feet. There were some good suggestions and
    references, many thanks to all contributors. Most of the responses agreed
    that placing the marker on the centre of rotation would yield accurate
    information, whilst this might be difficult in practice, in particular
    trying to find the ankle joint centre for the three foot designs.

    Regards,

    Tam
    ======================================

    1.

    Dear Tam, thank you for your interesting topic. I have no established

    answers, but perhaps the same interest, so I will look forward to your

    summary. Rather, I may suggest that this is a perfect topic ... at the

    Congress we are organising here in Bologna. You might have come across to

    the i-FAB initiative and the relevant Congress (see in case www.i-FAB.org).

    In addition to the topics addressed at the scientific sessions, meeting will

    be dedicated to consensus activities, and tutorials are being designed to

    meet perhaps these interests. Your topic is particularly appropriate also

    because of the presence and co-chair of people from INAL in Budrio

    (basically our National Insurance Institute for Occupational Diseases,

    taking care of thousands of amputees every year). Do not hesitate for any

    recommandation / suggestion from your side.

    Kind regards,



    ************************************************** ***********************

    Alberto Leardini, DPhil

    Movement Analysis Laboratory

    Director Prof. Sandro Giannini

    Centro di Ricerca Codivilla-Putti

    Istituti Ortopedici Rizzoli

    Via di Barbiano 1/10, 40136 Bologna ITALY

    tel: +39 051 6366522

    fax: +39 051 6366561

    email: leardini@ior.it

    http://www.ior.it/movlab/





    2. Dear Tam,

    If you have not come across the following paper yet, I think you may
    find it helpful:

    Prince, F., Winter, D. A., Sjonnensen, G., & Wheeldon, R. K. (1994). A
    new technique for the calculation of the energy stored, dissipated,
    and recovered in different ankle-foot prostheses. IEEE Transactions
    on Rehabilitation Engineering, 2(4), 247-254.

    This paper discusses some problems associated with modeling the ankle
    complex on the prosthetic limb and presents an alternative approach to
    modeling the ankle and foot of the prosthetic limb other than simply
    matching the marker placement on the intact limb.

    Jeremy

    --
    Jeremy D. Smith
    School of Sport & Exercise Science
    University of Northern Colorado
    Greeley, CO 80639
    Phone: 970.351.1761 Fax: 970.351.1762
    email: Jeremy.Smith@unco.edu



    3. Dear Tam

    I do not consider 3DGA ankle marker to be a crucial factor.

    I view complete prosthetic foot as an inverted offset "T" into
    consideration.

    Hence an adjustable pyramid (ankle) location may ease the alignment
    process.

    The below mentioned points should be considered in selecting a prosthetic
    foot

    1) Location of pyramid (ankle) with respect to length of foot.

    2) Length of foot with reference to sound foot in case of unilateral BK
    amputees.

    3) Stiffness of fore foot as well heel damper.

    Answers to your question mentioned below.

    1) ( NOT CRITICAL)

    2) ( ADJUSTABLE PYRAMID ANKLE TOWARDS FORE OR REAR FOOT

    ADVISED)

    3) (CERTAINLY RIGHT WITHOUT IGNORING OTHER POINTS)

    Mr.Rabinder Sahni

    Prosthetics R&D, Designer lower limbs, Self user

    INDIA


    4. Tam,
    There does not appear to be a standard marker placement for foot-ankle
    motion in prosthetics. Most mimic the contralateral limb in unilateral
    amputees, although I have never understood the logic there. The goal is to
    accurately capture the motion of the prosthesis, and presupposing anatomical
    symmetry where it clearly does not exist could confound that goal.

    Your suggestion to place the marker at the ankle joint center is a good
    idea, but difficult in practice. The SACH foot you plan to test and many
    dynamic feet have a solid ankle and a deformable leaf spring. Finding the
    center of rotation of a joint is fairly straightforward, and even if we
    allow that the instantaneous center of rotation changes, our errors will be
    small. However, finding the ICR of a beam deforming in bending is far more
    complex. Add to that the fact that the cosmetic cover of a prosthetic foot
    deforms as well, and the traditional placement of a marker at the ankle
    joint axis becomes less attractive.

    A better approach, though it can still have problems, is to place a marker
    triad on the dorsum of the prosthetic foot. This permits 3-D tracking of the
    foot segment independent of any shared ankle marker.

    The marker triad method is described here (and elsewhere, undoubtedly):
    Siegel KL, Kepple TM, Caldwell GE. Improved agreement of foot segmental
    power and rate of energy change during gait: inclusion of distal power terms
    and use of three-dimensional models. Journal of Biomechanics
    1996;29(6):823-827.

    More information on errors associated with the shared ankle approach in
    prosthetics is here:
    Geil MD, Parnianpour M, Quesada PM, Berme N, Simon SR. Comparison of methods
    for the calculation of energy storage and return in a dynamic elastic
    response prosthesis. Journal of Biomechanics 2000;33(12):1745-1750.

    Regards,
    Mark Geil



    > > > >

    Mark D. Geil, Ph.D.
    Associate Professor and Director, Biomechanics Laboratory
    Department of Kinesiology and Health, Georgia State University
    P.O. Box 3975
    Atlanta, GA 30302-3975
    NEW Phone: (404) 413-8379
    Email: mgeil@gsu.edu



    5. Tam,
    I would answer your questions in the reverse order:
    3-yes, placing the marker at the axis of the prosthetic foot would be more
    accurate for measuring the dorsiflexion/plantarflexion angle of the given
    foot.
    2-As a prosthetist, we TRY to align the prosthesis such that it mimics
    the unaffected side as best as possible. This is not always possible as
    feet are made in only select sizes (length and width) and the heights are
    not adjustable. Placing the marker at the level of the unaffected malleolus
    would appear to give better knee angle measurements of symmetry, but any
    initial flexion built into the prosthesis would negate the accuracy.
    And 1- I have not seen any 'standard protocal' for marker placement on the
    prosthetic side versus the unaffected side.
    I would advise you to specify your placement on EACH foot type, including
    rationel for the placement. And to chart the level of the marker off of the
    floor for each individual foot, as mentioned above, not all feet have the
    same height: so their axis will be in different places.

    Good luck, I look forward to reading your results.

    Kyle Sherk, CPO/LO
    Oklahoma City, Ok, USA
    ====================================


    On Mon, Feb 11, 2008 at 2:11 PM, Tam Nguyen wrote:

    > Dear all,
    >
    > We are about to begin a study looking at the functional performance
    > between three prosthetic feet (SACH, dynamic and multi-axial)using 3D gait
    > analysis in our VICON equipped laboratory. We found the following articles
    > with description of the ankle marker placement as on the spot corresponding
    > to the lateral malleolus on the unaffected side.
    >
    > * Sjodahl et al. Kinematic and kinetic gait analysis in the sagittal plane
    > of trans-femoral amputees before and after special gait re-education.
    > Prosthet Orthot Int 2002; 26: 101-112
    >
    > * Bateni and Olney. Effect of weight of prosthetic components on the gait
    > of tran-tibial amputees. JPO 2004; 16(4):113-120
    >
    >
    > We also found a few similar articles but the authors did not specify the
    > location of the ankle maker placement. We want to make sure that maker
    > placements are appropriately placed to represent the function of the feet
    > design and to avoid potential errors in the kinematics and kinetics
    > measurements. So our questions are:
    >
    >
    > 1. Is there a standard maker placement for prosthetic feet study?
    >
    > 2. What would be the reasons for choosing the same spot to the lateral
    > malleolus on the unaffected side?
    >
    > 3. Would it be more accurate to place the maker where the approximate
    > ankle joint centre/ axis of rotation is located? (As these locations are
    > definitely different within the three feet designs and certainly are
    > different from the unaffected side)
    >
    > Look forward to your responses and will post a summary.
    >