    Dear Biomec-L subscriber:

    A summary of the answers sent for the subject about "EMG and electric
    stimulation" is below.

    Thank you very much for:
    - Young-Hui Chang
    - Vera Talis
    - Edmund Cramp
    - Duncan Wood
    - Anand Pandyan
    - Paul Taylor
    - Colleen Monaghan
    - Kevin Kelleher - kevin.kelleher@strath.ac.uk
    - Wayne Piekarski - Piekarski@worldviz.com

    Best regards,

    Wagner de Godoy
    Gait Analysis Laboratory
    AACD - Brazil



    ----------

    Original question:

    I would like to know if to use an electric stimulation devices (FES and
    Russian current) and to acquire EMG signals simultaneously is possible.
    - is it possible to filter the interference of stimulation device?
    - does some rehabilitation center develop researches with this configuration
    of equipments?

    ----------

    Summary:


    I do not believe that there would be a safety issue but, as with any new
    procedure, I would start with a pilot test on a normal volunteer.

    Assuming that the frequency used by the Russian Current device is in the
    2.5kHz range then the EMG system should be able to filter it out - unless
    the Russian Current signal is large enough to interfere with the common mode
    circuitry in the preamplifier. This means that the Russian Current applied
    to the preamplifier inputs would need to be less than 4V ... higher levels
    will not damage the preamplifier but would interfere with the CMRR circuitry
    and cause interference - this interference would be impossible to filter
    out.

    Edmund Cramp
    eac@motion-labs.com


    ----------

    We tried to collect EMG during FES, but because of many cross-talks the EMG
    signal was corrupted.

    Vera Talis
    veratalis@yahoo.com

    ----------

    It depends if you want to close the loop and use the recorded EMG to drive
    the FES or if you want to only record EMG and filter the data off-line.

    Filtering off-line to measure voluntary activity is a simpler problem
    (relatively speaking). We recently published a short technical letter on how
    to do this using an optimum eigen filter (let me know if you want a
    reprint): Yeom HJ, Park YC, and Chang YH. Eigen filter to detect volitional
    EMG signals in autogenic EMG-controlled FES. Electronics Letters 43(25):
    1410-1411, 2007.

    A real-time, closed loop system is much harder, particularly with surface
    EMG recording and stim. A common method to cancel the primary stim artifact
    is to use a blanking circuit during the time of
    stimulation (i.e., don't record when stim is on). However, this still leaves
    residual stim artifacts along with the muscle response to the stim (M-wave),
    which can cause problematic false negatives due to poor SNR.

    A postdoc in my lab (Hojun Yeom) has developed a real-time closed loop
    system for surface electrode FES using blanking circuits and an adaptive
    filter that works reliably in real-time. We are currently
    writing these results up, but he published the original concept paper back
    in 2005 (Yeom et al. Gram-Schmidt M-Wave cancellor for EMG controlled FES.
    IEICE Trans. Inf. & Syst. E88-D(9): 2213-2217).

    A few groups have combined blanking circuits with comb filters with some
    success. But, comb filters are finite impulse response (FIR) filters and
    assume the artifacts to be deterministic signals, which
    we have found to be more problematic compared to the adaptive filter
    solution.

    Young-Hui Chang
    yh.chang@ap.gatech.edu

    ----------

    Quick note
    take a trigger fro FES and only colllect EMG when FES is off - I think TMSi
    and/or Delsys can do this

    Anand Pandyan
    a.d.pandyan@shar.keele.ac.uk

    ----------

    What stimulation parameters will you use?

    I used a 50Hz burst of 15 pulses from a stimulator made in our university
    department (University of Twente). Then measured EMG with a TMI
    International PortiLab EMG device. As you know, the stimulation will
    generate stimulation artefact in your EMG signal, there is a paper on
    stimulation artefact removal by O'Keefe et al. We made a Matlab file based
    on their algorithm. Check out our article "Interaction of Artificial and
    Physiological Activation of the Gastrocnemius During Gait. Monaghan, Colleen
    C.; Hermens, Hermie J.; Nene, Anand V.; Tenniglo, Martin J. B.; Veltink,
    Peter H. Neuromodulation, Volume 11, Number 2, April 2008 , pp. 135-142(8)".
    This paper has details of the EMG equipment we used as well as containing
    the reference of the paper for stimulus artefact removal and how we did it.
    Hope this is useful.

    Colleen Monaghan
    c.c.monaghan@gmail.com

    ----------

    It is possible to detect EMG while using electrical stimulation but
    generally requires some method of blanking and filtering. Please find below
    a reference list which was completed in 2004.

    6. Kraft GH, Fitts SS and Hammond, MC. Techniques to improve function of
    the arm and hand in chronic hemiplegia. Arch. Phys. Med. Rehabil. 1992; 73:
    220-227.

    7. G Francisco, J Chae H Chawla, S Kirshblum, R Zorowitz, G Lewis and S
    Pang. Electromyogram-triggerred neuromuscular stimulation for improving the
    arm function of acute stroke survivors: a randomised pilot study. Arch Phys
    Med Rehabil 1998.Vol. 79

    8. J Heckmann, T Mokrusch, A Krockel, S Warnke, T von Stockert and B
    Neundorfer. EMG-triggered electrical muscle stimulation in the treatment of
    central hemiparesis after stroke. Eur j phys med rehabil 1997; 7 No 5
    pp138-142

    12. HJ Hollander, M Huber, G Vossius. An EMG controlled multichannel
    stimulator. Advances in external control of human extremities IX Ed. Dejan
    Popovic. Belgrade 1987 pp 291 - 295

    13. J. Minzly, J. Mizrahi, N. Hakim, A. Liberson. Stimulation artefact
    suppression for EMG recording during FES by a constant-current stimulator.
    Med. & Biol. Eng. & Comput., 1993 31, 72-75

    14. Timothy RD Scott, Kevin L Kilgore, and P Hunter Peckham. Assessment of
    tri-state myoelectric control for bilateral upper extremity neuroprostheses.
    Proc. 5th Vienna international Workshop on Functional Electrical
    Stimulation. Aug 17-19 1995 ISBN 3-900928-03-7 p 343 - 346

    15. S Saxena, S Nikolic, D Popovic An EMG-controlled grasping system for
    tetraplegics. J.Rehabil Res & Dev Vol.32 No.1 1995

    16. Thoresen R, Ferrarin R, Spadone R, Frigo C. An approach using Wrist
    extension as control of FES for restoration of hand function in
    tetraplegics. Proc. 6th Vienna international Workshop on Functional
    Electrical Stimulation. Sep 22-24 1998 ISBN 3-900928-04-5 p 263 - 266

    17. S Sennels, F Biering-Sorensen, OT Andersen and SD Hansen. Functional
    neuromuscular stimulation controlled by surface elctromyographic signals
    produced by volitional activation of the same
    muscle: adaptive removal of the muscle response from recorded EMG signal.
    IEEE Trans Rehabil Eng Vol 5 No 2 June 1997

    19. Muraoka Y, Miyajima S, Tomita Y, Hona S, Tanaka N, Okajima Y.
    EMG-Controlled hand opening for hemiplegia. Proc. 6th Vienna international
    Workshop on Functional Electrical Stimulation. Sep 22-24
    1998 ISBN 3-900928-04-5 p 255-258.

    20. Rakos M, Freudenschub B, Grirsch W, Hofer C, Kaus J, Meiners T,
    Paternostro T, Mayr W. EMG controlled FES treatment of the paralysed upper
    extremity. Proc. 6th Vienna international Workshop on Functional Electrical
    Stimulation. Sep 22-24 1998 ISBN 3-900928-04-5 p 259 - 262

    21. Tepavac D, Medri E. Programmable functional electrical stimulator with
    EMG feedback. Proc. 6th Vienna international Workshop on Functional
    Electrical Stimulation. Sep 22-24 1998 ISBN 3-900928-04-5 p 267 - 270

    22. Kimberley TJ, Lewis SM, Auerbach EJ, Dorsey LL, Lojovich JM, Carey JR.
    Electrical stimulation driving functional improvements and cortical changes
    in subjects with stroke. Exp Brain Res. 2004
    Feb;154(4):450-60

    23. Cauraugh J. Light K, Kim S. Thigpen M, BehramA. Chronic motor
    dysfunction after stroke: recovering wrist and finger extension by
    electromyography triggered neuromuscular stimulation. Stroke. 2000
    Jun;31(6):1360-4

    29. Thoresen R, Spadne R, Ferrarin M. A pilot study of myolectrically
    controlled FES of upper extremity. IEEE Trans Neural Syst Rehabil Eng. 2001
    Jun;9(2):161-8

    30. Thoresen R, Occhi E, Brccardi S, Ferrarin M. Tenonesis grip augmented by
    EMG controlled FES. 9th IFESS conference, Bournemouth Sept. 2004. Accepted
    for publication.

    31. Muraoka Y. Development of an EMG recording device from stimulation
    electrodes for functional electrical stimulation. Front Med Biol Eng.
    2002;11(4);323-33

    32. Rakos M, Freudenschuss B, Girsch W, Hofer C, Kaus J, Meiners T,
    Peternostro T, Mayr W. Electromyogram-controlled functional electrical
    stimulation for treatment of the paralysed upper extremity. Artif Organs.
    1999 May;23(5): 466-9

    33. Popovic MB, Popovi DB, Sinkjaer T, Stefanovic A, Schwirtlich L.
    Restoration of reaching and grasping promoted by functional electrical
    therapy. Artif Organs. 2002 Mar;26(3):271-5

    34. Chae J, Bethoux F, Bohine T, Dobos L, Davis T, Friedl A. Neuromuscular
    Stimulation for Upper Extremity Motor and Functional Recovery in Acute
    Hemiplegia. Stroke. 1998; 29:975-979.

    36. Chae J, Fang ZP, Walker M Purmehdi S. Intramuscular elecromyographicaly
    controlled neuromuscular electrical stimulation for upper limb recovery in
    chronic hemiplegia. Am J Phys Rehabi 2001;80:935-941

    37. Van Overeem Hansen G. EMG controlled Functional Electrical Stimulation
    of the paretic hand Scand J. Rehab Med 1979.11:189-193

    38. Taylor P, Chappell P. Variation in system gain when using voluntary EMG
    to control electrical stimulation of the same muscle. 9th Annual Conference
    of the International FES Society and 2nd
    FESnet Conference, (ISBN 1-85899-191-9), pp. 126-128, Bournemouth, UK,
    September 2004.



    Paul Taylor
    p.taylor@mpbe-sdh.demon.co.uk

    Duncan Wood
    d.wood@mpbe-sdh.demon.co.uk


    ----------