    Dear Netters,

    A long time ago I posted the following two questions. I'm sorry
    about the delay in posting my summary of responses. I'd like to
    take this chance to thank everyone who helped me with these
    problems.

    thanks again
    Pat Fanning

    ----------------------

    >Dear Netters,
    >
    >A friend is using Fuji prescale pressure measuring film to measure to
    >pressure in bone joints. The film he is using is designed to measure
    >Super Low pressure, in the range of 5 to 25 kgf/cm2. The film changes
    >colour in accordance to the amount of pressure applied to it. I have
    >two questions about this product.
    >
    >1) Does anyone know how to convert kgf/cm2 to Pascals?
    >
    >2) Has anyone any experience in translating these or similar patterns
    > into pressure data? We have access to a colour scanner, so does
    > anyone know of any programs that can divide a colour image into
    > regions of similar colour?
    >
    >Thanks in advance for any help you can provide. As usual I will provide
    >a summary of answers received.
    >
    > Pat Fanning

    ----------------------

    The answer to the first question is fairly obvious.

    1 kgf = 1 * g N =~ 9.81 N

    => 1 kgf/cm2 = 10000kgf/m2 = 98100 N/m2 = 98.1 kPa.

    Thanks to all those who replied to this question. They were

    Hans Vos (j.e.vos@med.rug.nl)
    Dieter Rosenbaum (diro@sirius.medizin.uni-ulm.de)
    Graham Robinson (mecgcr@clust.heriot-watt.ac.uk)
    Dan Levine (dll@zim.bms.com)
    Jesus Dapena (dapena@valeri.hper.indiana.edu)
    Brian W. Bergemann (BWBERGEMANN@ussa-admin.ussa.edu)
    Eric Powell (epowell@fs1.ho.man.ac.uk)
    Joseph Hale (jeh7a@galen.med.virginia.edu)
    Carlos Melendez (carlos@pliers.unm.edu)
    Christian Haid (Christian.Haid@uibk.ac.at)
    Thomas G. Loebig (tom@biomechanics.asri.edu)
    Christopher Allen Favre (Favre@aol.com)
    Mike=Gross%PT%MAH@css.unc.edu
    Orit YARDEN (yarden@eng.tau.ac.il)
    Phil Morrison (PHILM@tekotago.ac.nz)
    Peter M. Quesada (peter@gait1.gait.ohio-state.edu)
    Moshe Nissan (nissanm@techunix.technion.ac.il)
    Charles Luevano (chuck@slider.unm.edu)
    E. C. "Ned" Frederick (72735.77@compuserve.com)

    ----------------------

    The following is a list of replies to the second question.

    ----------------------

    From: diro@sirius.medizin.uni-ulm.de (Dieter Rosenbaum)

    Pat:

    2. We have been using a digital image analysis system which -after calibration-
    provided the information on contact area, average and peak pressure (for
    measurements of the Chopart joint loading before/after calcaneal fractures).
    The product of contact area and average pressure will give you an estimate
    of the transitted force. I don`t know how to use the scanner for that
    application but it will probably depend mostly on the capabilities of the
    software you are using and the resolution of the scanner.

    I hope this helped, feel free to contact me if you have additional questions.
    Ciao,

    Dieter...

    ************************************************** *****************************
    * Dr. Dieter Rosenbaum voice0)731 - 502 3492 *
    * Abteilung Unfallchirurgische fax: (0)731 - 502 3498 *
    * Forschung und Biomechanik email: diro@sirius.medizin.uni-ulm.de *
    * Universitaet Ulm *
    * Helmholtzstr. 14 *
    * 89081 ULM *
    * Germany *
    ************************************************** *****************************

    ----------------------

    From: "Joseph E. Hale"

    >
    >2) Has anyone any experience in translating these or similar patterns
    > into pressure data? We have access to a colour scanner, so does
    > anyone know of any programs that can divide a colour image into
    > regions of similar colour?
    >

    I have do a fair amount of work with Pressensor film, both in terms of
    characterization its behavior and application to joint contact stress
    measurement. As a starting point, you make want to look at the following
    papers:

    Singerman, RJ, Pedersen, DR and Brown TD: Quantitation of Pressure
    Sensitive Film Using Digital Image Scanning. Experimental Mechanics
    27:99-105, 1987.

    Hale, JE and Brown, TD: Contact Stress Gradient Detection Limits of
    Pressensor Film. Journal of Biomechanical Engineering 114:352-357, 1992.

    Most image analysis software packages will allow you to look at image
    intensity (which can be correlated to applied pressure) for individual
    pixels within the region of interest. We used a commercial package from
    Precision Visuals (PV-Wave) on a DEC workstation. Another option for the
    Macintosh is a package called Image, which is available free through the NIH.

    I hope this information is helpful. Please feel free to contact me directly
    if you have additional questions.

    Joe Hale

    Director, Orthopaedic Biomechanics Laboratory Tel: 804-924-5989
    University of Virginia FAX: 804-924-1691
    Charlottesville, VA USA email: jhale@virginia.edu

    ----------------------

    From: "J.B. FINLAY"

    Dear Pat:

    In quantifying your Fuji-film stains, you may wish to refer to
    the following papers which address the applications and technical
    limitations of this material.

    In particular, you should be wary of assigning a "meaningless"
    number of Pressure Ranges to your images. The Pressure Ranges
    should be established with respect to the standard deviations
    achieved when Calibration Stains are produced at nominal
    pressures. The number of "valid" Pressure Ranges with such a
    quantitative approach may be small; however, "image averaging"
    (using several images) can improve the resolution and increase
    the number of meaningful ranges.

    Liggins, A.B. and Finlay, J.B. Recording contact areas and
    pressures in joint interfaces. In Experimental Mechanics (Ed E.
    G. Little), 1992, pp. 71-88 (Elsevier Science Publishers,
    Amsterdam).

    Liggins, A.B., Stranart, J.C.E., Finlay, J.B. and Rorabeck,
    C.H. Calibration and manipulation of data from Fuji
    pressure-sensitive film. In Experimental Mechanics (Ed E. G.
    Little), 1992, pp. 61-70 (Elsevier Science Publishers,
    Amsterdam).

    Liggins, A.B. and Finlay, J.B. Sterilization of Fuji pressure-
    sensitive film. Medical Engineering & Physics 16(?): In Press,
    May 1994.

    Liggins, A.B., Hardie, W.R. and Finlay, J.B. The spatial and
    pressure resolution of Fuji pressure-sensitive film.
    Experimental Mechanics, In Press, November 1994.

    Liggins, A.B. and Finlay, J.B. Image-addition provides enhanced
    resolution in analyses with Fuji pressure-sensitive film.
    Experimental Mechanics, 1994, Submitted.

    Liggins, A.B., Surry, K. and Finlay, J.B. Sealing Fuji
    Prescale pressure-sensitive film for protection against fluid
    damage: The effect on its response". Strain: Submitted, July
    1994.


    For further information or reprints of these papers, you should
    contact my colleague
    Adrian Liggins, PhD at:

    519-663-3843 Phone
    519-663-3904 FAX

    Liggins@ortholab.uh.rri.uwo.ca

    His e-mail is currently out-of-service, so any e-mail to him can
    be directed through my account.

    Best wishes:

    Bryan Finlay, PhD 519-663-3063
    Director of Orthopaedic Research 519-663-3904 FAX
    University Hospital
    P.O. Box 5339
    London, Ontario, CANADA, N6A 5A5

    ----------------------

    From: "Thomas G. Loebig"

    My colleagues here have considerable experience with Fuji and will no doubt
    reply to you next week as they are out of town until Monday. If you haven't
    heard from Don Anderson or Tom Daniel soon, write back and I'll tell them of
    your query.

    Good Luck

    Tom

    Thomas G. Loebig, MSME Research Associate
    Allegheny-Singer Research Institute tom@biomechanics.asri.edu
    320 E. North Avenue,10th Floor ST voice: (412)359-6773
    Pittsburgh, PA 15212-4772 fax: (412)359-3494

    ----------------------

    From: Favre@aol.com

    Pat,
    As far as analysis of the pressure data. I have had some experience
    with data similar to this; however it was topographic. The principle though
    is still the same. I would enter this data in as a matrix. This would give
    the pressure data on a 2D plane. With this data you could enter it into a
    program like Mathematica, MathCAD, or even the HP48's to get a contour or
    surface plot. Mathematica is really the best when it comes to the surface
    plots. The surface plot will give you a 3D rendering of your data. You must
    note though that this plot is dependant upon how the film was placed. For
    example, if the film was flat and a force was placed upon it, the x,y axis
    would be rectangular. If it was placed around a spherical surface, for
    example a ball and socket joint, you must take into account the curvature.
    Simple vector calculus can yield how the grid lines must then be drawn. I
    have a program in FORTRAN and C++ which can analyze the data and come up with
    the peaks, and their location. This is only good though for flat surfaces.
    When it comes to simplifing the data, I have used Harvard Graphics Plus.
    It has a mechanism that can read adjacent colors, and simplify them. This
    is similar to what happens on TV when they hide a person's face. The main
    problem that I have encountered though is with the scanner. If you use an
    inferior scanner, it may read and put colors that are not really there, and
    thus distort the scanned image.
    I hope that these ideas will help your friend. If you would like a copy
    of the peak program, just write me.

    -Christopher Allen Favre
    Sophmore, Misssissippi Gulf Coast Community College

    ----------------------

    From: PHILM@tekotago.ac.nz

    Pat, I have had similar problems with analysis of thermographic images. The
    most basic solution is relatively easy in principle since you only have to
    be count each pixel of the scanned data. Unfortunately I have only found
    very expensive options to the dilemma and time pressure prevents me from
    writing the software myself.

    The one of the cheapest options so far has been image processing tools
    developed for the package LabView. It runs on the Mac/IBM/Sun and the tools
    were developed by a French firm who's address I cannot lay my hands on
    at the moment (cost US$1200 to 3200). However the US address is
    GTFS Inc., 2455 Bennett Rd, #100C, Santa Rosa, CA95404

    Two other sources were
    Image Analyst by Automatix, 755 Middlesex Turnpike, Billerica, MA US$2000
    (again a little too sophisticated for my purposes)
    Enhance v2.x by Microfrontier, 7650 Hickman Rds, Des Moines, IA50322. (US$375)
    Note both only run on the Macintosh

    I would appreciate if you come across any other solutions. This problem is on
    a back burner for me at the moment.
    My immediate solution is to visit either the Pathology/ Anatomy/ Physiology
    departments and use theirs. A case of a sledge hammer to crack a nut.

    Best of luck

    Phil Morrison
    New Zealand

    ----------------------

    From: hound-dog

    as far as using your color scanner, i think you might be able to to it.
    what you would have to do is to calibrate it. if you could apply known
    pressures, for instance with a materials testing machine, you could scan
    those in and use for references as your calibration.

    from the work with what i have done with the same film, i do believe it
    works by popping more bubbles (with the developer for the film) at
    higher pressures. if you do a close up on a low pressure region and
    count the corpuscles which have popped and developed, i think you will
    find them less per unit area than for a larger pressure

    hope this is of help.

    ************************************************** ******************************
    Matt Rupert rupertmp@ucunix.san.uc.edu
    Department of Aerospace Engineering
    and Engineering Mechanics
    University of Cincinnati, Ohio
    ************************************************** ******************************

    ----------------------

    From: Matt

    Prof. Roger Haut of Michigan State University (East Lansing, Michigan, USA)
    has used Prescale film in measurements of cartilage loading during patella
    impact in rabbits. Try the following email address:

    haut@msm.msu.edu

    or call US phone number: 517-355-0320

    ************************

    Matthew Reed
    Senior Research Associate
    University of Michigan
    Transportation Research Institute
    matt@biosci.umtri.umich.edu

    ************************

    ----------------------

    From: "E. C. Frederick"

    We have had some limited expreience with Fuji film. In our experience it is
    best used as a measure of relative pressure distribution. Incidentally I
    believe Fuji sells a scanner with analysis software to convert the various
    colors to pressures.

    In my experience 5 to 10 kgf/cm^2 pressure is not all that low for a
    biomechanical system. Plantar pressures, in walking for example, can peak in
    that range.

    __________________________________________________ __________________________
    E. C."Ned" Frederick, Ph.D. Phone: 603.772.2505
    Exeter Research, Inc. Fax: 603.772.5463
    8 Chestnut Street
    Exeter, NH 03833-1859 USA EMail: 72735.77@Compuserve.com
    __________________________________________________ ___________________________

    ----------------------

    From: dll@pendragon.zim.bms.com (Dan Levine)

    Hello Pat:

    Re question 2) above: Our Tribology group does not utilize tools which
    distinguish colour variations in Fuji images. They use some simple technique
    to measure the overall area only.

    Dan Levine
    Zimmer, Inc. / Warsaw, Indiana, USA
    dll@pendragon.zim.bms.com