    Dear Netters:

    A few days ago I posted the request below. I would like to thank
    you all for such quick and informative responses---especially over
    a holiday weekend, (US & Canada). I would like to particularly thank
    Dr. David Dean of Case Western Reserve University for making available
    a copy of his in-press lit. review on the topic.

    Thanks again to all.

    -Bob

    And while I have your divided attention:
    re: The resent Russian yenta. As I have stated before, I am all in favour
    of Flame-ing GROSS offenders of nettequete (send them a copy of the World
    Almanac if it makes you feel better). I also think it's great to inform the
    offender's POSTMASTER (omitting the Almanac of course). But, why perpetually
    re-injure the List (and eat up bandwidth) by re-including the original post
    in a List's broadcast mailing. If we've subscribe, we've seen. While I would
    like to send out kudos to Mitchell Gil Maltenfort for finding the current
    offender's true address, I was dismayed to find two instances of the post in
    the latest issue of the (biomch-L) list (not to mention the three other
    versions I encountered in other Lists' members' complaints.)
    Just 2 cents... I now return you to my original intent...

    Original post:

    {snip}
    >Some time ago I recall reading an ad for a laser based measurement/modeling
    >device. The device operated by placing the object to be measured on a
    >platform, a laser then scanned across the surface and reflectance(?) or
    >phase shift(?) was used to obtain the measurements. In this way a model of
    >the object could be obtained and exported to various software packages.
    {snip}

    REPLYS {trimmed}~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~ ~~~

    David Dean Ph.D.
    Lewis L. Sadler and
    Louis-Philippe Amiot ....................................
    Each sent a copy of David Dean's extremely useful survey. I am including
    the LEAST abbridged version here for those with a slightly different need.
    It is long, but I believe well worth it.

    +---------------------------------------------------------------------+
    | David Dean, Ph.D., Assistant Professor |
    | Departments of Anatomy, Orthodontics & Biomedical Engineering |
    | Bolton-Brush Growth Study Center, 3080 Bolton Dental Bldg |
    | Case Western Reserve University Voice: (216) 368-1975 |
    | 10900 Euclid Avenue FAX: (216) 368-3204 |
    | Cleveland, OH 44106-4905 USA E-mail: dean@lucifer.cwru.edu|
    |WWW: http://www.cwru.edu/CWRU/Dept/Dent/orth/ortho/dean/deanpage.html|
    +---------------------------------------------------------------------+

    In Press (LA Marcus, ed.): Advances in Morphometrics. NYC:Plenum

    3D Data Capture and Visualization
    David Dean
    Departments of Anatomy, Orthodontics, and Biomedical Engineering
    Case Western Reserve University
    10900 Euclid Avenue
    Cleveland, OH 44106-4905 U.S.A.
    E-mail: dean@lucifer.cwru.edu

    Data Capture Devices I: Hand digitization

    Electromagnetic Devices

    1.1 DigiTracker ($3395)
    Visual Circuits
    3309 83rd Avenue North
    Brooklyn Park, MN 55443
    tel (612) 560-6205
    fax (612) 659-6629

    As of this writing the device has not yet been released. An OEM
    version of the Ascension Flock of Birds, it is reported to be a
    work-alike of the Polhemus 3DRAW. Like the Flock of Birds it's
    accuracy rating is a magnitude lower, 0.1" RMS. It may have
    greater utility for biometric applications than an unaltered
    Flock of Birds device, which Visual Circuits retails for $2575,
    because of its well-designed stylus. It is supplied with a
    powerful digitizing software environment, BitBlitzer. A number of
    mainstream Draw, Video, and CAD software export filters are
    supplied with BitBlitzer.

    1.2 Flock of Birds ($2695)
    Ascension Technology Corp.
    P.O. Box 527
    Burlington, VT 05402
    tel (802) 860-6440
    fax (802) 860-6439

    An early split-off from the Polhemus group. The same technology
    has been adapted to tracking multiple electromagnetic 6D (xyz
    coordinates and normals) sensors ("a flock") within the magnetic
    field emitted by a single stationary source. Popular device for
    biomechanical studies of joint movement. Translational accuracy
    rated to 0.1" RMS. Above price is for single sensor device.

    1.3 3DRAW ($6500)
    Polhemus
    One Hercules Drive, P.O. Box 560
    Colchester, VT 05446
    tel (800) 357-4777/(802) 655-3159
    fax (801) 655-1439

    This fourth generation device has been used by several
    investigators to study non-human primate and human crania (see
    Dean, this volume; Lele and Richtsmeier, 1992; Corner and
    Richtsmeier, 1992; Richtsmeier and Walker, 1993; Vannier and
    Conroy, 1989). Originally developed to be mounted on a fighter
    pilot helmet to track head movements (Raab et al., 1979), this
    application was soon obsoleted by laser-based sighting devices.
    It consists of a platform under which a fixed 3 axis magnetic
    source is emitted. A specimen can be fixed to the surface of this
    platform. This object is digitized with the aid of a hand-held
    stylus.
    The stylus is equipped with a 3 axis magnetic sensor that
    allows a small computer to determine the location of the tip in 3
    space, including xyz normal information. Because it assumes the
    magnetic field is constant, the work area should be kept
    reasonably clear of ferrous objects. None should be within a
    meter of the digitizing envelope, and other large iron objects
    such as steam heated radiators and filing cabinets should be 2-3
    meters away. Given a changing ferrous environment, the device
    should be recalibrated (it comes with calibration software)
    before each digitizing session. Unless the stylus is used very
    carefully, the normal information is not very important as part
    of the morphometric data collected. However, the viewpoint of the
    data collection interface can be tied to the stylus. This allows
    easy perusal of the data from various points of view, a crucial
    error-checking step whenever potentially overlapping coordinates
    are collected.
    Their is firmware control of the stylus data report; it can
    be set in point or continuous mode on the fly. Continuous mode is
    very useful for tracing edge features (see Dean, this volume).
    The accuracy of this device has been tested internally by
    Polhemus using widely accepted methods (Krieg et al., 1992) and
    has shown accuracy of 0.01" RMS (Root-Mean-Square), or 0.254 mm.
    While this is more than a two-fold improvement over the original
    (first generation) 3SPACE device, the generation of the device
    that has commonly been used with large non-human primate and
    human skull material, it is still below the acceptable range for
    most dental studies.
    The latest version of this device, the 3DRAW, is highly
    portable. The 3DRAW table, stylus, and data handling device all
    together weigh less than 12 pounds. There is a padded, hard-
    shell, airline check-in carrying case available.
    An interface was recently released that allows the data
    collected by this device to be sent to the CADKEYS generic CAD
    program. This flexible PC program can be readily tailored to most
    biometric applications. It is included, gratis, with the device.
    Another longstanding interface for Polhemus, Metrecom, Scientific
    Accessories (see the latter two below) devices is marketed by
    Mira Imaging (2257 South East, Suite 1A, Salt Lake City, UT
    84106, tel. 801/466-4641, fax. 801/466-4699). This interface,
    HyperspaceTM, is primarily aimed at producing a surface mesh of
    solid objects. The NATO morphometrics workshop attendees tested
    this software and found it difficult, but not impossible, to
    adapt to morphometric applications such as single point or space
    curve digitization.

    Laser Light Stylus-based Digitizers

    1.4 Flashpoint Model 3000i ($14,900)
    Pixsys, Inc.
    5680-B Central Avenue
    Boulder, CO 80301
    tel (303) 447-0248
    fax (303) 447-3905

    A device like this or the much more expensive Optotrak (section
    1.4) is likely to be optimal for many morphometric studies. These
    devices do not have the echo or humidity artefacts of acoustic
    digitizers or the artefacts due to changing magnetic fields
    common with electromagnetic digitizers. These devices are,
    technologically, the newest, thus their high cost.

    Both devices consist of one or more LEDs (Light Emitting Diodes)
    or fiber optic light emitters. These light sources are detected
    by an array of 3 light sensors mounted in a bracket so that they
    are fixed in a non-collinear position. This fixed offset is used
    to triangulate the position of the light source. The light source
    can be attached to a stylus. Data can be collected in point or
    continuous mode. There are device drivers for all the common CAD
    programs.

    The rated accuracy of the current model is 0.5 mm. Over the next
    few years I expect to see rapid increases in the accuracy of this
    device coinciding with price reductions. The entire device
    weights 22.4 pounds and should be quite portable.

    1.5 Optotrak ($57,500)
    Northern Digital, Inc.
    403 Albert Street
    Waterloo, ON N2L 3V2, Canada
    tel (800) 265-2741/(519) 884-5142
    fax (519) 884-5142

    This device has basically similar components and works in the
    same fashion as the Pixsys. Its accuracy is rated at 0.1 mm in
    the x-y plane, 0.15 mm in z, when the stylus is within 2.25m of
    the sensors. Weighing approximately 90 pounds, not including a
    computer to record and visualize data, it is not easily
    transported.

    Acoustic Stylus-based Digitizer
    1.6 GP-12 ES ($3200)
    Science Accessories Corporation
    2 Research Drive, P.O. Box 825
    Shelton, CT 06484
    tel (203) 925-1661
    fax (203) 929-9636

    The stylus in this device produces a clicking sound. This sound
    is differentially detected by three wall-mounted acoustic
    sensors. With the three separate readings of the click, the
    location of the source can be triangulated. The stylus should be
    at least as flexible as that of the 3DRAW or DigiTracker (I have
    not tried to use it), making complex concavities more reachable
    by this probe than, for example, those used in fixed-base, servo-
    mechanism digitizing arms (see next section). However this
    maneuverability is not without problems. Convexities may,
    however, prevent sound from reaching the acoustic sensors without
    distortion. High humidity is also known to effect the quality of
    the data. The manufacturer's customer's claim an accuracy of
    0.01", 0.254 mm., in a cubic meter digitizing envelope. There is
    a real-time driver that works with a large number of CAD programs
    including: AutoCAD, Cadkey, Mastercam. Scientific Accessories is
    also willing to develop a customized interface.

    Rigid and Servo-mechanism Arms
    1.7 Cyclone ($52,000)
    Renishaw, Inc.
    623 Cooper Court
    Schaumburg, IL 60173
    tel (708) 843-3666
    fax (708) 843-1744

    This device comes right out of the rapid prototyping industry. It
    includes a computer terminal and digitizing base. This unjointed
    arm digitizer sits in place, retreating in the Z direction as
    protrusions on the object are passed under its probe on a
    linearly moving platform. Successive cross sections are collected
    in this manner to build up an isosurface of the object. It moves
    slowly and gently enough during its slow traverse that there
    should be little chance for damage of specimens (stiff casts
    would be at less risk than friable originals). However, complex
    convexities are ignored. A similar device, the Retroscan
    ($20,000) uses a drill bit to mill a model of a digitized object
    from solid plastic blocks.

    1.8 Diagraph 202 ($1472)
    Seritex Inc.
    450 Barell Avenue
    Carlstadt, NJ 07072
    tel (201) 939-4606
    fax (201) 939-3468

    The Diagraph, designed by Swiss manufacturer GPM, is used to
    directly digitize objects (e.g., a bone) held rigidly in place.
    One end of its rigid arm points at surface features while the
    opposite end of the arm puts a pen on paper. The paper sits on a
    horizontal plane of tracing (or graph) paper set at a particular
    z-height. The Diagraph is useful for the collection of serial
    sections for 3D reconstruction.

    1.9 Faro Arm ($14,400)
    Faro
    125 Technology Park
    Lake Mary, FL 32746
    tel (800) 736-6063/(407) 333-9911
    fax (407) 333-4181

    Morphometricians have attempted to use digitizing arms for at
    least the past 30 years. It has not been a successful in most
    craniofacial applications because the objects of study have been
    too topologically complex. Both of these devices have four
    joints, giving them a great deal of flexibility. Even so it might
    be difficult capturing some points in the orbits, temporal fossa,
    posterior choanae, the palette and parts of the basicranium.
    Tracing lines on bony tori or elsewhere on the skull is likely to
    range between difficult and impossible. The accuracy of the
    Bronze Series 6 is 0.012 inches, 0.3 mm, within a 6 foot
    spherical diameter.

    Faro manufactures a more expensive line of digitizing arms, the
    Surgicom series. These are used in human surgical planning,
    execution, and follow-up. A complete neurosurgical package,
    including visualization software, has recently been approved for
    use by the FDA is marketed by ISG Technologies (6509 Airport
    Road, Mississauga, ON L4V 1S7, Canada, tel. 905/672-2100, fax.
    905/672-2307). The digitizing arm is primarily used as a virtual-
    reality tool in designing a surgical plan, digitizing landmarks
    to set up a reference frame on the patient that matches that in
    the surgical plan, and as a position tracker for surgical
    instruments. Advances in the use of this and other technologies
    for frameless stereotactic surgery (Zinreich et al., 1993;
    Sandeman et al., 1992) will likely be furthered by the
    incorporation of newer morphometric techniques.

    1.10 Romer Model 2000 ($105,000)
    Romer Supratech, Inc.
    5145 Avenida Encinas
    Carlsbad, CA 92008
    tel (619) 438-7802
    fax (619) 431-7940

    This model is the smallest of a line of 4 arms with digitizing
    envelopes of between 6.5' and 9.8' in diameter. These arms have
    rotary joints making them more maneuverable than the Faro arm.
    These devices are also designed to be portable. The Model 2000
    weighs 17 pounds. The small base includes a table clamp and an
    i/o processing unit with a serial port. It can thus be easily
    adapted to use with a portable computer. The accuracy rating is
    0.00125" RMS or 0.032 mm as tested by the NIST (National
    Institute of Standards and Technology).

    Data Capture Devices II: Remote Digitization
    In this section we will not discuss multisurface, slice-based
    approaches such as CT or MR-scanning. The two types of digitizers
    discussed here, laser light and optical, are primarily used to
    digitize one surface, the external surface. Optical systems
    calibrate the X and Y values from the power of magnifcation and
    the Z can be determined by the focus level.

    Laser light digitization has a reasonable possibility of
    providing data that can be used to construct highly accurate
    surface representations of the original object, i.e., an
    isosurface. Point resolution is a function of the width of the
    light source beam. Scan accuracy is determined not only by beam
    width, but also camera sensitivity, surface shininess, and the
    system's ability to screen out artefactual reflections. In most
    cases users will benefit from importing the data into a
    visualization package and digitizing the required 3D landmarks
    and/or lines there.

    Multiple Point Source Laser Light Scanning

    2.1 Digibot II ($49,000)
    Digibotics
    2800 Longhorn Boulevard, Suite 102
    Austin, TX 78758
    tel (512) 832-6544
    fax (512) 832-1163

    The Digibot II uses a four-axis laser light scanner which checks
    sequential points on the surface. This allows some correction for
    artifacts. The four axis array also obviates focusing problems
    that other systems may have.

    2.2 Hirez Model 3030 ($46,400)
    Cyberware Laboratory, Inc.
    2110 Del Monte Avenue
    Monteray, CA 93940
    tel (408) 675-1440
    fax (408) 675-1494

    This device consists of a moving source of laser light that is
    detected by a video camera at a known angle of offset. Like all
    of this first group of laser light digitizers, the Cyberware
    scanners use multiple laser light sources. This allows the device
    to cover more surface area in less time. An entire human face can
    be scanned in less than 10 seconds. This is regularly done for a
    number of different morphometric applications (Cutting et al.,
    1988; Wohlers, 1992). However, because of the diffuse light
    source artifacts, especially of complex surfaces, tend to corrupt
    the data. The recommendation of scanning a surface from different
    orientations and co-registering the scans can lead to compounded
    artifacts. Assuming no artifacts, landmark accuracy is
    approximately 0.02 inches, 0.5 mm.

    One version of this device also captures color information from
    the surface which can be used to render a more realistic
    isosurface. One version of this device weighs less than 50
    pounds.

    2.3 Hyscan ($n.a.)
    Hymarc Ltd. (Ottawa, Canada)
    G.A. Davis Associates, Inc.
    P.O. Box 36240
    19959 Vernier Avenue, Suite 2B
    Harper Woods, MI 48225
    tel (313) 886-4101
    fax (313) 886-1107

    A multiple point laser device with very high accuracy, +/-
    0.001", and relatively slow data acquisition on large objects.
    The Hyscan accumulates data at the rate of 10,000 points per
    second.

    2.4 Optica ($50,000)
    3D Technology, Inc.
    12 Cambridge Drive
    Trumbull, CT 06611-4764
    tel (203) 371-8500
    fax (203) 371-6300

    Using a technology developed in England, but reminiscent of the
    Cyberware scanner, this vendor claims 0.0008", 0.02mm accuracy
    (Vinarub and Kapoor, 1992).

    2.5 Rapid Profile Sensor ($40,000)
    Laser Design, Inc.
    9401 James Avenue South, Suite 162
    Minneapolis, MN 55431
    tel (612) 884-9648
    fax (612) 884-9653

    This device is scheduled to be available shortly. The developers
    of this device claim digitizing speeds comparable with the
    Cyberware scanner with greatly improved accuracy. A whole body
    version of this scanner is scheduled to ship at $145,000.
    Single Point Source Laser Light Scanning

    2.6 CyberScan ($30,000)
    CyberOptics Corp.
    2505 Kennedy Street NE
    Minneapolis, MN 55413
    tel (612) 331-5702
    fax (612) 331-3826

    Sharnoa Corp.
    45901 5 Mile Road
    Plymouth, MI 48170
    tel (313) 925-1661
    fax (313) 454-7198

    This single source laser sits at a pre-set "stand-off" distance
    from the surface to be digitized. It has a high accuracy, 0.001"
    (0.0254 mm.), but can fail on surfaces with convexities that
    block the path of the scan head. These devices would take several
    hours to scan an entire human face. Sharnoa uses the same scan
    head and provides integration with CNC machine milling tools for
    rapid prototyping.

    2.7 Laser Triangulation Probe LTP 60 ($28,700 scan head, $200,000 complete
    device)
    Carl Zeiss, Inc.
    One Zeiss Drive
    Thomwood, NY 10594
    tel (914) 747-1800/681-7849
    fax (914) 681-7454

    Perhaps the most accurate digitizing head on the market, the
    accuracy of this device is rated at 2.0 microns, 0.0008". The
    probe collects 100 coordinates per second when positioned 115 mm
    away from the object of interest. An observation beam tracks the
    surface to maintain this standoff distance. The device is
    supplied with a visualization environment, Holos-UX, which allows
    the user to define and locate coordinates of landmarks and
    surface patches on the digitized object.




    ============== Begin part 2 ==========================

    2.8 Lasar and TriCam ($50,000 and $11,000, respectively)
    Perceptron, Inc.
    23855 Research Drive
    Farmington Hills, MI 48335
    tel (810) 478-7710
    fax (810) 478-7059

    Single laser point devices that use radar focusing. The
    digitizing envelope of the Lasar is commonly a cube of 2-3 meters
    with an accuracy of 0.1 mm. The TriCam is more accurate in its
    smaller digitizing envelope. The TriCam weighs a portable 20
    pounds.

    2.9 MessTechnik OMS 400/250 ($40,000 scan head, $193,000 complete device)
    Wegu Inc.
    Building II
    Orchard Ridge Corporate Park
    Fields Lane
    Brewster, NY 10509
    tel (914) 277-5753
    fax (914) 277-5830

    A close second to the Zeiss device, the accuracy of this device
    is rated to 2.2 microns, 0.000098". The rate of data capture is
    5000 points per second. The full device uses a video source and
    tactile probe to insure it is following the surface carefully.
    The standoff distance from the object is a function of lens
    magnification, from 15 to 75 mm.

    2.10 OP5 ($30,000)
    Renishaw, Inc.
    623 Cooper Court
    Schaumburg, IL 60173
    tel (708) 843-3666
    fax (708) 843-1744

    This laser scanning probe has a rated accuracy of 0.0001". The
    probe rotates about a single axis up to 120 degrees, but must be
    within 14mm of the scanned surface. It collects 200 points per
    second.

    2.11 Surveyor ($70,000)
    Laser Design, Inc.
    9401 James Avenue South, Suite 162
    Minneapolis, MN 55431
    tel (612) 884-9648
    fax (612) 884-9653

    The developers of this device claim digitizing speeds comparable
    with the Cyberware scanner and accuracy improvements of at least
    a magnitude, i.e., 0.05 mm. This device would be optimal for
    studies requiring "dental" precision levels and rapid data
    acquistion.

    Moire
    2.12 CadEyes ($80,000)
    Medar Company
    38700 Grand River
    Farmington Hills, MI 48335
    tel (810) 477-3900
    fax (810) 477-8897

    As with all moire photography, light is passed through a screen
    grating. The resulting topographic pattern can be used for 3D
    digitization. Medar claims 0.0001", 0.00254 mm, accuracy in the Z
    dimension with this device.

    2.13 Mini-Moire ($39,000)
    Electro-Optical Information Systems
    528 Euclid Street
    Santa Monica, CA 90402
    tel (310) 451-8566
    fax (310) 393-2453

    Electro-Optical Information Systems device works on the same
    principle as the CadEyes scanner. It claims about the same
    accuracy level, however it is much more portable.

    Convergent Photogrammetry

    2.14 Photomodeler ($895)
    (Windows)
    Eos Systems, Inc.
    2040 West 12th Avenue
    Vancouver, BC V6J 2G2 CANADA
    tel (604) 732-6658
    fax (604) 732-4716

    This software uses standard photographs or 35mm slides to produce
    a precise 3D model (AutoCAD 3D *.DXF file) of any object. The
    documentation shows you how to calibrate your camera's focal
    length, image aspect ratio, image position, and lens distortion
    for this application. There are a few more planning few steps to
    take so that all views will be complementary.

    ~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~ ~~~~~~~
    From: j-patton@nwu.edu (Jim Patton)

    Try BROWN & SHARPE in Providence, RI. They make laser scan devices.

    JIM PATTON, Northwestern U. (j-patton@nwu.edu)

    ~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~ ~~~~~~~
    Sender: grerfx00@ccmail.ca.boeing.com

    I think what you're looking for is a company called
    Cyberware. Here is their address:

    2110 Del Monte Avenue
    Monterey, CA 93949-3797
    phone: 408-657-1450
    fax: 408-657-1494

    Good luck.

    Rush Green
    grerfx00@ccmail.iasl.ca.boeing.com

    ~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~ ~~~~~~~
    Sender: d.vashishth@qmw.ac.uk

    Try contacting Zygo Corporation, Laurel Brook road, Middlefield, Conn. USA.
    They make laser telemeter system which measures dimensions of object and is
    based on the phase-shift principle.

    Deepak Vashishth

    ~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~ ~~~~~~~
    Sender: rkippenc@vax7.curtin.edu.au

    Curtin University of Technology have access to a 3-D scanner. It is Sunday,
    so I do not have access to any other information but will furnish you with
    more details if you think it is what you are after. Jill Smith is the
    computer artist and Phil Dench the software engineer. The system was used
    on Jarassic Park, although Phil is/has developed software
    (Cysurf-Cyberware)). The scanner can laser scan a bust in 15 seconds and
    the image immediately appears on the screen, consisting of 250,000
    geometrical points. The image can be rotated, have the colour stripped out
    or have portions selected for cutting pasting into other programs. Jill
    uses it for milling heads and other objects.

    Cameron Kippen

    ~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~ ~~~~~~~
    Sender: dwyer@conn.me.queensu.ca

    A company you might want to call would be Hymarc in Ottawa, Ontario. We
    recently had them scan a series of total knee femoral components and we
    have had fantastic results with this data. They are a relatively new
    company that has made major advances in this area. Their system seems to
    be able to scan almost any shape from the joystick of a fighter jet to
    the panels of cars. We imported our data and were able to create 3D
    solid models. Good luck!

    Kimberly A. Dwyer, M.S.

    ~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~ ~~~~~
    Sender: ukorde01@ype.gmpt.gmeds.com

    Such a device is marketed by a firm called
    LASER DESIGN Inc. from Minneapolis, Minnesota.
    I know of it being used for reverse engineering
    products and also for measurement/inpsection
    applications.

    -Uday

    ~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~~ ~~~~~~~
    Sender: cgdanias.pg@swansea.ac.uk

    The following company in the U.K. have laser insturments which
    measure distances. I am using one of their lasers to measure
    small displacements on teeth and I reckon that they probably have
    insturments which will do what you require.

    The company is,
    GRAHAM & WHITE INSTURMENTS LTD.
    135 Hatfield Road
    St.Albans
    Hertfordshire
    AL1 4LZ
    ENGLAND

    Tel. (code to UK) 01727 59373
    Fax. (code to UK) 01727 44272

    Christo