    ORIGINAL QUESTION:

    I was interested in finding more information on current trends in materials
    and design of total joints (hip, knee). Specifically, I read an article by
    D. Dowson (1990) that proposed a "cushion" type joint that could come
    closer in reproducing the elastohydrodynamic mode of lubrication that
    occurs in normal synovial joints. Current designs rely on boundry
    lubrication alone which affords higher levels of friction resulting in
    premature wear and loosening of the prosthesis. I found the ideas very
    interesting but did not know if further research had produced any tangible
    results.

    ******************************************
    SUMMARY OF RESPONSES:

    Date: Thu, 31 Aug 1995 20:17:21 +0000
    From: clfr50 To: alhkg@ttuhsc.edu
    Subject: Re: biomaterials
    X-Sun-Charset: US-ASCII


    ->TI: FRICTIONAL CHARACTERIZATION OF EXPLANTED CHARNLEY HIP PROSTHESES
    AU: HALL_RM, UNSWORTH_A, WROBLEWSKI_BM, BURGESS_IC NA: UNIV DURHAM,CTR
    BIOMED ENGN,DURHAM DH1 3HP,ENGLAND
    UNIV DURHAM,CTR BIOMED ENGN,DURHAM DH1 3HP,ENGLAND WRIGHTINGTON HOSP,CTR
    HIP SURG,WIGAN,ENGLAND JN: WEAR, 1994, Vol.175, No.1-2, pp.159-166 IS:
    0043-1648
    AB: The frictional resistances of 54 explanted and 5 new Charnley hip
    prostheses were measured using one of two Durham hip function simulators.
    The friction factor f was measured in both dry and lubricated environments
    up to a viscosity eta of approximately 0.1 Pa s. The friction factor values
    for the new prostheses were found to be 0.11 (+/-0.025) and 0.04 (+/-0.01)
    for dry and lubricated (eta=0.01 Pa s) regimes respectively. The
    distributions of f for the new prostheses were found to be approximately
    Gaussian. However, the distributions for the explanted prostheses were
    skewed, with most retaining friction values similar to those of new
    prostheses, but a significant proportion exhibiting increased frictional
    resistance. For dry regimes 30% of the prostheses showed values of f in
    excess of 0.16, while for lubricated regimes (eta=0.01 Pa s) 39% showed
    values of f greater than 0.07. All the joints indicated a mixed lubrication
    regime when analysed in terms of a Stribeck plot. A weak negative
    correlation (P