    Dear BIOMCH-L readers,

    While the item below is somewhat outside the target domain of our list (as
    Biomechanics and Kinesiology are still very much a research endeavour, with
    limited confidence on the part of many physicians), we all hope that our work
    will be of benefit in public health matters. Therefore, I think that the sad
    history of the Therac-25 Radiation Therapy Machine warrants dissemination over
    our list as much as over the other lists mentioned below.

    Regards -- Herman J. Woltring (List owner/editor, BIOMCH-L@HEARN)

    [EARN/BITNET/NETNORTH users can subscribe to the Risks Digest at the listserver
    LISTSERV@FINHUTC via the SUBSCRIBE RISKS "full_personal_name" command. Each
    week, about two issues are distributed with contributions from individuals
    world-wide.]

    ---------------------------------------------------------------------------

    Distribution-File:
    health-l@irlearn !International Discussion on Health Research
    !healthco@rpicicge !Communication in Health/Medical Context
    medinf-l@dearn !Medical Informatics Discussion List
    oncogene@irlearn !BIOSCI Oncogenes Bulletin Board
    radsig@uwavm !Radiology Special Interest Group
    support@rpicicge !Moderators for Healthco

    RISKS-LIST: RISKS-FORUM Digest Wednesday 11 January 1989 Volume 8 : Issue 5

    FORUM ON RISKS TO THE PUBLIC IN COMPUTERS AND RELATED SYSTEMS
    ACM Committee on Computers and Public Policy, Peter G. Neumann, moderator

    (........)

    Date: Tue, 10 Jan 89 12:04 EST
    From: RMorris@DOCKMASTER.ARPA
    Subject: Medical software

    A Digital Matter of Life and Death
    by Ivars Peterson
    Science News, 12 March 1988

    The radiation-therapy machine, a Therac 25 linear accelerator, was designed to
    send a penetrating X-ray or electron beam deep into a cancer patient's body to
    destroy embedded tumors without injuring skin tissue. But in three separate
    instances in 1985 and 1986, the machine failed. Instead of delivering a safe
    level of radiation, the Therac 25 administered a dose that was more than 100
    times larger then the typical treatment dose. Two patients died and a third
    was severely burned.
    The malfunction was caused by an error in the computer program controlling
    the machine. It was a subtle error that no one had picked up during the
    extensive testing the machine had undergone. The error surfaced only when a
    technician happened to use a specific, unusual combination of keystrokes to
    instruct the machine.
    The Therac incidents and other cases of medical device failures caused by
    computer errors have focused attention on the increasingly important role
    played by computers in medical applications. Computers or machines with
    built-in microprocessors perform functions that range from keeping track of
    patients to diagnosing ailments and providing treatments.
    "The impact of computers on medical care and the medical community is the
    most significant factor that we have to face," says Frank E. Samuel Jr.,
    president of the Health Industry Manufacturers Association (HIMA), based in
    Washington, D.C. "Health care will change more dramatically in the next 10
    years because of software-driven products than for any other single cause."
    Samuel made his remarks as a recent HIMA-sponsored conference on the regulation
    of medical software.
    At the same time, reports of medical devices with computer-related
    problems are appearing more and more frequently. In 1985, the Food and Drug
    Administration (FDA) reported that recalls of medical devices because of
    computer faults had roughly doubled over the previous five years. Since then,
    the number of such complaints has risen further.
    The problems range across a wide spectrum of computer-based medical
    devices. A system designed for monitoring several patients at once was
    recalled because it kept mixing up the patients. A programmable heart
    pacemaker suddenly "froze" while it was being adjusted by a doctor. A device
    for dispensing insulin delivered the drug at an inappropriate rate. An expert
    system gave the wrong diagnosis, resulting in a patient receiving a drug
    overdose. An ultrasound scanner sometimes underestimated fetal weight.
    "No one can deny that allowing computers to perform some of the functions
    normally carried out by trained and licensed medical professionals raises
    questions concerning the personal health and safety of citizens," Michael
    Gemignani of the University of Maine in Orono comments in ABACUS (Vol. 5, No.
    1). "But even if we agree something more needs to be done to protect society
    in the face of these technological innovations, we are still left with the
    question: What should be done and by whom?"
    The FDA, in its mandated role as guardian of public health and safety, is
    now preparing to regulate the software component of medical devices. The
    agency's effort has already raised questions about what kinds of products,
    software and information systems should be regulated.
    Last fall, the FDA published a draft policy for the regulation of computer
    products marked for medical use. In that policy, the concept of "competent
    human intervention" sets the dividing line between what is and is not
    regulated. In other words, the computer product in question is subject to
    regulation if a qualified doctor or nurse cannot effectively intervene to
    override the machine's actions. Devices such as software-driven cancer therapy
    machines, programmable heart pacemakers and automatic drug dispensers clearly
    fall into that category.
    On the other hand, the FDA states that it would not regulate computer
    products that simply store, retrieve and disseminate information analogous to
    that traditionally provided by textbooks and journals. In addition, the
    agency's regulations would not apply to computer products used only for
    communications, general accounting or teaching.
    For example, a physician may use a computer program known as an expert
    system to help make a diagnosis. Because the expert system does not directly
    drive another medical device that, say, could dispense a drug when needed, and
    because the doctor can make an independent judgment, such an expert system
    would be exempt from FDA rules governing medical devices.
    However, the greatest advantage of software - its flexibility - is also,
    from a regulatory point of view, one of its biggest problems. Computer
    programs are easy to change and can be used in many different ways. If
    corrections are made or new features added, how much scrutiny should the
    modified version of a previously approved computer product undergo? That
    question is still unresolved.
    Furthermore, it's sometimes hard to make a clear distinction between
    programs that perform a "library" function and those that can be classified as
    being part of a medical device. A case in point is the patient medical record,
    traditionally a file folder containing various sheets of paper listing
    treatments, medical observations and other pieces of information vital for the
    patient's proper care.
    Many hospitals are now moving toward medical records that are stored on a
    computer. The difficulty arises when such information systems are connected
    directly to machines that, for example, record patient blood pressure and heart
    rate. If a nurse takes down the data and then enters the figures into a
    computer, the information system software would not be subject to FDA rules.
    But if the machine sends the data directly to the computer, then the
    information system is considered by the FDA to be an "accessory" to a medical
    device and subject the same level of regulation as the machine itself.
    Information system vendors disagree with the FDA's position. They argue
    that the FDA does not presently have rules governing the quality and content of
    paper medical records. There's no reason for the FDA to start regulating such
    records, they say, just because the records happen to be in a computer's memory
    rather than on paper. In fact, using a computer-based system would
    dramatically reduce the incidence of errors in patient records, the vendors
    claim. The benefits of improved record keeping would clearly outweigh the need
    for burdensome regulation.
    The FDA's James S. Benson concedes that "regulation is not the automatic
    solution to problems in hospitals and elsewhere." Nevertheless, the agency
    must comply with a 1976 law that contains a broad definition of what
    constitutes a medical device. Interpreted in its broadest sense, the
    definition encompasses practically everything used in a hospital, from X-ray
    machines to pencils.
    FDA officials say they recognize the difficulties involved in regulating
    medical software. "The agency fully appreciates the revolution occurring in
    medicine with the introduction of computers and microprocessors," says Frank E.
    Young, FDA commissioner. "We're taking a reasoned, structured approach with a
    minimum of oversight. We have tried to give general guidelines. The policy
    has been deliberately made flexible."
    The flexibility allows the FDA to consider applications for approval on a
    case-by-case basis. That limits the "chilling fear of undue regulation," says
    Young. Furthermore, as technologies change and experience with computers in
    medical applications grows, decisions on how much regulation is needed may also
    change.
    To many manufacturers and users of medical products, the FDA's idea of
    flexibility leaves too much uncertainty and opens up the possibility of
    increased regulation in the future. "The FDA casts too wide a net," says
    Edward M. Basile of King & Spalding, a law firm in Washington, D.C. "Their
    basic assumption is that everything should be regulated."
    "There's no disagreement about the extremes," says Harold M. Schoolman of
    the National Library of Medicine in Bethesda, Md. "The question is how and
    where to draw the line between the extremes." The important issue, he says, is
    maintaining a balance between appropriate safeguards and incentives for
    innovation.
    Even in situations where it's clear that certain software ought to be
    reviewed, the FDA faces the additional difficulty of how to go about verifying
    that a particular computer program does what it's supposed to do -- nothing
    more, nothing less. As experience with software for other applications has
    shown, the task of checking software quality can be overwhelming (SN: 9/13/86,
    p. 171).
    A few years ago, when most medical devices did not contain computers, it
    was relatively easy to foresee all possible inputs and to check the
    consequences of each one, says James Howard of General Electric's Medical
    Systems Group in Milwaukee, Wis. With computers, the number of possible paths
    is greatly increased. "It's more important than ever to build safe products
    that perform as required," he says. But because a detailed analysis takes so
    long, it often can't be done. "This is a major concern to both manufacturers
    and the FDA," says Howard.
    The FDA defines software as a "set of instructions that enables a
    computing machine to control, monitor or otherwise interact with a medical
    device." The proposed regulations require a software developer to show that the
    algorithm, or mathematical recipe, used in the computer program is appropriate
    and has been implemented correctly in the software. The FDA also requires
    assurance that any software failure would not injure the patient.
    How that assurance can be provided is still unclear. Techniques for
    evaluating software safety are relatively new. Who does the checking, how much
    evidence is enough and whether the FDA can perform an independent check are
    also unresolved issues. Furthermore, software developers are wary of
    submitting complete listings of the instructions in their computer programs
    because competitors may get a look at this "source code" by making a request to
    the FDA under the Freedom of Information Act.
    The trouble with the FDA approach, says Howard, is that it doesn't
    consider under what conditions software is used. Instead, the FDA ought to
    focus on the idea that not all computer errors are equally serious. Using a
    kind of hazard analysis to focus on situations that could lead to
    life-threating computer failures would be one way to eliminate the most serious
    potential faults and to shorten testing times.
    Software developers also need to improve the methods they use for
    constructing computer programs. We need to "industrialize" software
    development so that programs are written in a consistent way, says James
    Dobbins of Verilog USA, Inc., in Alexandria, Va. Too often, programmers
    include a description of what each part of a program does only as an
    afterthought. They rarely go back to clean up or polish a program to make it
    more understandable.
    Software development can be standardized and automated, says Dobbins.
    "The tools are there to industrialize the whole process. You just have to go
    find them."
    Programmers, on the other hand, complain that they're in a no-win
    situation. Software is continually modified as it evolves, often to meet
    demands for new features to make the product more competitive. In the rush to
    market, when delays can put a company at a competitive disadvantage, software
    testing often loses out. Delays in completing a software package are balanced
    against the possibility of failing to root out potentially embarrassing errors.
    This is the kind of situation that can lead to lawsuits, says Vincent
    Brannigan, an attorney in Adelphi, Md. Software is clearly a product, he says.
    If it's defective and injures a consumer, then the manufacturer is liable.
    Among the faults Brannigan lists is the tendency of software and computer
    companies to promise more than they can fulfill and to cut costs by doing less
    testing. This is the only field, he says, where the customer is expected to
    pay for finishing a product through the purchase of periodic updates and
    corrections to the software.
    "Disclaimers don't mean anything," Brannigan says. "The product should
    have been right in the first place." That means paying much more attention to
    how software is written and tested. "The software must look as shiny and clean
    as the rest of the machine," he says.
    So far, software developers have generally escaped damaging lawsuits and
    settlements, but that may change. To many medical-device producers, the threat
    of litigation may be even more effective than proposed FDA regulations for
    assuring the quality of products.
    Even finding out what went wrong is a time-consuming process. The FDA and
    other groups are still investigating aspects of why the Therac 25, manufactured
    by Atomic Energy of Canada Ltd. in Kanata, Ontario, failed. What's evident is
    that the problem could probably have been avoided if an appropriate safety
    analysis had been done.
    The Therac 25 delivers two forms of radiation: either a high-energy
    electron beam or, when a metal target intercepts the electron beam, a
    lower-energy X-ray beam. It turns out that when a nimble, experienced
    technician punches in a particular sequence of commands faster than the
    programmers had anticipated, the metal target fails to swing into place.
    A safety analysis would have identified the missing target as a
    potentially dangerous situation. The machine could have been programmed so
    that it couldn't operate if the target, as confirmed by a sensor, were not in
    place.
    Perhaps such a complex, computer-driven machine wasn't even necessary. By
    sacrificing a little convenience and flexibility, a machine with a simple
    on-off switch and a timer could probably have done the same job - with a much
    smaller chance of failure.

    [This is a familiar topic to RISKS readers, but this particular article
    is extremely well written and seems worth including, even if old.
    (RISKS has reported one additional death involving the Therac.) PGN]

    ------------------------------